Measuring laser output with Lux meter
Measuring laser output with a lux meter is possible but requires careful conversion between photometric and radiometric units. The lux meter's frequency response graph indicates the sensitivity of the sensor, which is crucial for accurate measurements. For a 532 nm green laser, the luminous efficacy is around 0.86, meaning a 20 mW laser would yield approximately 17 lumens. To achieve a reading of 4000 lux, the detector area must be about 0.004 m², which is larger than the lux meter's 40 mm aperture. While reasonable accuracy can be achieved with a lux meter, an optical power meter is recommended for more precise measurements.
jodyflorian
Hiya,
Question 1: (Main question)
I bought a lux meter while understanding very little about how light is measured. I'm hoping to measure the output of a laser in mW. I assume it's possible to do this at least to an accuracy of about 10-20% which is all I need? If so, please could someone help me out with the equation?
The lux meter came with a frequency response graph, so for example if I measure a green laser (532nm) I know that I need to reduce the reading to 95%... But I'm not sure if this is a graph for the sensitivity of the human eye for various frequencies, or if it's the sensor's sensitivity for frequencies. If it's the latter, then I would need a function for the human eye's sensitivity to frequencies to calculate the power of a laser as well!
Thanks!
PS: I've shone a supposed 20mw laser onto the lux meter (green, 532nm) and it reads around 4000 lux - so this the ballpark figure for double checking the equation :)
Science news on Phys.org
1 Lux = 1 Lumen/m^2 and is a photometric unit. Watts (or mW) is a radiometric unit. So, not only do you need to convert a radiometric unit to a photometric unit (using the "luminous efficacy" curve), but also you need to convert a measurement of incidance to flux.
I highly recommend you get some reading material to familiarize yourself with the subject: NIST has some online material, I think, but for me the gold-standard reference is "Introduction to Radiometry" By William Wolfe.
Unfortunately, you should have obtained an optical power meter, not a luminometer for this measurement (unless there is a specific reason why you need to measure things photometrically)
In any case, let's run through your numbers: 4000 lux and 20 mW of 532 nm laser light. The luminous efficacy curve gives a value near 1 (0.86) at that wavelength, so the effective "eyeball watts" is 20*0.86 = 17 lm. In order to convert lm to lux, you need to specify the size of the beam or size of the detector: in order to get 4000 lux from 17 lm, I calculate a detector that is 0.004 m^2 or 6.5 cm on a side: large, but reasonable.
Does that help?
Thanks Andy!
I think I follow, or at least I think I will once I have a good look over it lol...
So do you think I could get some fairly reasonable accuracy by using the lux meter?
If it helps to know, the lux meter has a translucent circular aperture with diameter 40mm, and the laser beam is approx 2-3mm diameter.
Your explanation has already helped me follow the maths of it all but if you can, I'd be very grateful if you could clarify firstly what you meant by '6.5 cm on a side', (i.e. if the 40mm aperture is relevant?) and if you think I need to buy an optical power meter or if the lux meter should do the job?
Thanks!
10K