- #36
Tom.G
Science Advisor
Gold Member
- 5,487
- 4,306
The "LOAD VOLTS" graph in post #14 shows an error 0.07%! You can get that much variation just by wiggling an alligator clip on a contact, and I doubt your meter is that accurate. If the meter has 4-digit readout that is the ±1 digit uncertainty inherent in all digital measurements.
For the graph in post #1 RCALC vs RMEAS, two possibilities come to mind:
You mentioned that you used the same meter to measure the current. The Current range of a digital ammeter has some low, non-zero, internal resistance, which would add to your effective load resistance. Has this been taken into account?
Beg, borrow, or steal a second meter to measure the current at the same time as the voltage measurement.
Please let us know what you find.
Cheers,
Tom
For the graph in post #1 RCALC vs RMEAS, two possibilities come to mind:
- the meter (or its leads) need calibration and/or repair
- the battery holder is defective
You mentioned that you used the same meter to measure the current. The Current range of a digital ammeter has some low, non-zero, internal resistance, which would add to your effective load resistance. Has this been taken into account?
Beg, borrow, or steal a second meter to measure the current at the same time as the voltage measurement.
Please let us know what you find.
Cheers,
Tom