In summary: V = -5.00; // raw mV buffer1 = 0.00; // buffer 7.00 buffer2 = 177.48; // buffer 4.00 cal_mV1 = -15.00; // pH reading during calibration using 7.0 buffer (point 1) cal_mV2 = 160.0; // pH reading during calibration using 4.0 buffer (point 2) referenceRange = buffer2 - buffer1; readingRange = cal_mV
DeepThinking
Hi, I'm trying to calculate the slope of a pH probe using 7.0 and 10.0 buffer solutions. I've been searching for hours but all examples use 7.0 and 4.0. I've calibrated my probe using the 7.0 and 4.0 buffers and calculated the slope, then I used 7.0 and 10.0, used the same math for the slope and it was quite different so I don't know if that's normal or I'm doing something wrong.
This is using 7.0 and 4.0 and confident its correct.
pH electrode generated –5 mV in pH 7.00 buffer and +165 mV in pH 4.00 buffer.
165 mV – (-5 mV) = 170 mV
170mv/3 = 56.66mV/pH unit
56.66/59.16 x 100 = 95.8% slope
Now this is an example with 7.0 and 10.0, is this correct?
pH electrode generated –5 mV in pH 7.00 buffer and -165 mV in pH 10.00 buffer.
-165 mV – (-5 mV) = -160 mV
-160mv/3 = -53.33mV/pH unit
53.33/59.16 x 100 = -90.15% slope
Obviously with a 2nd buffer over 7.0 the slope always comes out as a negative number, would I just invert it to a positive number?
Thanks
Chemistry news on Phys.org
Your approach to calculating the ΔV is inconsistent. If you go for Vat lower pH - Vat higher pH you should do it for both cases.
Likes DeepThinking
Thanks, something like this?
This is using 7.0 and 10.0
-5 mV – (-165 mV) = 160 mV
160mv/3 = 53.33mV/pH unit
53.33/59.16 x 100 = 90.15% slope
Yes.
Any one or two pH buffers will not be enough to allow for finding a slope for either pH. You need titration data for the region containing the pH change value of each buffer. This is not just one or two points in each pH region. You might take a slope using pH 7 and pH 10, but this is not the intent of doing a titration curve.
Thanks Borek!
Sorry symbolipoint I don't understand that. The formula seems to be common for some meters, I got it from a Hanna manual but it only shows the first example.
Can I ask about a formula for temperature compensation? Best I found is 0.003 pH/ºC/pH but I don't know how to apply that.
I might by some chance miss what you meant. Your first post #1 seemed to be about slope in relation to what SHOULD be using or making a titration curve, but you asked for slopes(?) for pH 7 and pH 10 indicators. No mention about temperature there was given. One would be looking at titrant volume versus pH (or mv) measurement,and then use the points to find the slope at the inflection point, FOR EACH intended indicator solution.
I really do now think I misunderstood. Are you really looking for a slope? Calibration of the pH meter? If this, then you can use pH 7 AND pH 10 BUFFER SOLUTIONS. to ADJUST the pH meter. Someone else, for example, Borek, can explain this better.
Sorry for the confusion, I was only asking about the slope of the electrode using the mid and high calibration settings so I can check the condition of the electrode after calibration. What Borek said makes sense, I just wasn't sure.
Ultimately I'm trying to achieve the following.
1. Calculate slope of pH electrode using mid/low buffer, mid/high buffer or mid/low/high buffer. (I think this is figured out now)
2. Apply calibration using mid buffer (1 point), mid/low buffer (2 point), mid/high buffer (2 point) or
mid/low/high buffer (3 point).
3. Temperature compensation for calibrating and monitoring pH
For the 3 different types of calibration this is what I have, do the formulas look ok? Only 1 of the 3 would be used.
2 point calibration - 2nd point buffer is 4.0
mV = -5.00; // raw mV
readingRange = cal_mV2 - cal_mV1;
mV = (((mV - cal_mV1) * referenceRange) / readingRange) + buffer1;
pH = fabs(7.0 - (mV / 59.16));
2 point calibration - 2nd point buffer is 10.0
buffer1 = 0.0 mV // buffer 7.00
buffer2 = -177.48; // buffer 10.00
cal_mV1 = -15.0 mV // pH reading during calibration using 7.0 buffer (point 1)
cal_mV2 = -192.2 mV // pH reading during calibration using 10.0 buffer (point 2)
referenceRange = buffer1 - buffer2;
readingRange = cal_mV1 - cal_mV2;
3 point calibration
buffer3 = -177.48; // buffer 10.00
cal_mV3 = -194.48; // pH reading during calibration using 10.0 buffer (point 3)
if (mV < 0) // if pH greater than 7.0
referenceRange = buffer1 - buffer3;
readingRange = cal_mV1 - cal_mV3;
else // if pH less than 7.0
At this point I would have the calibrated pH value and would like to apply temperature compensation to it.
Thanks
IIRC, some clever pH meters happily self-calibrate their internal 'curve' for the whole range using but one buffer.
Even so, we checked ours daily at #4 & #10, to be sure, to be sure...
You must have a robust system in place for detecting and handling outages, such as a fouled electrode or stale buffer. I remember the scramble before we discovered that a certain pH-adjusted HPLC mobile phase would not play well with an electrode that wasn't 'squeaky clean'. As our routine daily checks had found that electrode was reading correctly, if a bit slower than usual, we were able to go back and double-check prior usage...
Also, we kept several sets of 'pH Strips' handy, as a sanity check...
Likes DeepThinking
Thanks Nik_2213, those are good tips to remember.
Related to Calculate the pH Slope using 7.0 and 10.0 buffers
What is the purpose of calculating the pH slope using 7.0 and 10.0 buffers?
The purpose of calculating the pH slope using 7.0 and 10.0 buffers is to determine the change in pH for a given change in concentration of a solution. This information is useful in understanding the acidity or basicity of a solution and can also be used to predict the behavior of a solution in different chemical reactions.
How do you calculate the pH slope using 7.0 and 10.0 buffers?
To calculate the pH slope, you first need to measure the pH of both the 7.0 and 10.0 buffers. Then, using the Henderson-Hasselbalch equation, you can calculate the ratio of the concentrations of the conjugate acid and base in each buffer. Finally, the pH slope can be determined by taking the difference between the pH values of the two buffers and dividing it by the difference in their logarithmic concentrations.
What factors can affect the accuracy of the pH slope calculation?
There are several factors that can affect the accuracy of the pH slope calculation. These include temperature, ionic strength, and the presence of other ions or substances that can interact with the buffer solutions. It is important to control these variables as much as possible to obtain reliable results.
What is the significance of the pH slope in chemistry?
The pH slope is significant in chemistry because it provides information about the strength of an acid or base and how it will behave in different reactions. It can also be used to determine the optimal pH for certain chemical processes and to monitor changes in pH during a reaction.
Can the pH slope be used to compare the acidity or basicity of different solutions?
Yes, the pH slope can be used to compare the acidity or basicity of different solutions. A steeper pH slope indicates a greater change in pH for a given change in concentration, suggesting a stronger acid or base. On the other hand, a flatter pH slope indicates a smaller change in pH, indicating a weaker acid or base.
9K
165K
10K
12K
31K