Can a Photomultiplier be Calibrated for Extremely Low Light Levels?

  • Thread starter MereA.
  • Start date
In summary, a photomultiplier is not a good option for measuring very low levels of light. There are other options available, such as avalanche photodiodes or photo-multipliers with gated pulses.
  • #1
MereA.
4
1
Does anyone know how to calibrate a photomultiplier? I need to measure very, very low levels of light (almost no light) and am considering using a photomultiplier. However, I don't know how easy (or difficult) it would be to calibrate a photomultiplier to be sensitive to such a low level of light. If a photomultiplier won't work, are there other options? I've looked at lux meters but could not find any that would measure below 10^-2 lux, which is too much light. Thank you for any help.
 
  • Like
Likes Prabhakar Gouripeddi
Engineering news on Phys.org
  • #2
Typically an Fe55 source which gives an x-ray photon of very consistent energy which will generate a known number of electrons in the photo cathode.

An simpler method is a pulsed LED source and a stack of ND filters
 
  • Like
Likes Prabhakar Gouripeddi
  • #3
Do you know if there is a good source (website, book) that would describe the process of calibration using an LED source and ND filters?
 
  • #4
In what physical quantity do you want it calibrated?
 
  • #5
If possible, I'd like to express the amount of light in terms of lux.
 
  • #6
You do know that the lux is a photometric unit of light intensity, right? As such, it is only valid for visible light and you have to find the SI definition of candela (through one standard source that I do not remember from the top of my head).
 
  • #7
Light is annoyingly difficult to calibrate accurately - better than the few % level.

Easiest way to do it is to compare a source with a light meter then use a stack of photographic ND filters to drop the source by a known amount for your low light sensor.
 
  • Like
Likes Prabhakar Gouripeddi
  • #8
Okay. Thank you for your help, NobodySpecial and Dickfore.
 
  • #9
2This url has a table that converts lux to photons.

http://www.egc.com/useful_info_lighting.php

For example, from the table, 54 sunlight photons (per m2 per sec?*) is equal to 1 lux. It is possible to calibrate the gain of a photomultiplier tube (PMT) by putting in a dark box, and by using a pulsed light source and neutral density filters (NDFs), get the light level so low that the distribution of number of photoelectrons from the PMT cathode is given by the Poisson distribution (i.e., less than 10 photoelectrons per light pulse) with discrete pulse heights. You will need a gated pulse height analyzer for this. By attenuating the light another factor of 10 using a NDF, nearly every output pulse is due to a single photoelectron. measuring the Coulombs per photoelectron is straight forward. For example, a single photoelecton in a 10-stage PMT might create about 10-13 Coulombs at the anode.

* Something is very fishy. 1 watt of sunlight per square meter is about 2.5 x 1018 photons per square meter, from basic physics.

[added] the table in the url is in mcromoles of photons per m2 per sec, or 6 x 1017 photons per m2 per sec. Using the conversion to watts in the table gives 4.57 x 6 x 1017 = 2.74 x 1018, in agreement with the above physics example.

Bob S
 
Last edited:
  • Like
Likes Prabhakar Gouripeddi
  • #10
MereA. said:
...If a photomultiplier won't work, are there other options?...

These devices can compete:
http://en.wikipedia.org/wiki/Avalanche_photodiode

You may also want to browse in our own selection of instruments, just in case.
www.gentec-eo.com (from fW to kW)
 
  • #11
APDs are generally easier than photo-multipliers if you don't need a very large area. They don't need scary HV supplies, have better red sensitivity and are pretty immune to damage from high light levels

The electronics for them is pretty simple - I had breadboard circuits doing individual photon counting with avalanche mode APDs 20years ago.
 
  • #12
Here is a good selection of Hamamatsu photomultipliers:

http://sales.hamamatsu.com/en/produ...s/photomultiplier-tubes/productlist.php?#form

As one example, the Hamamatsu R760 tube has a gain of about 1 x 106, and a dark current of about 1 nA, meaning the dark current is equivalent to about [STRIKE]600[/STRIKE] 6000 cathode photoelectrons per second. If you could gate the signal from the PMT, you could get a clean signal without a dark current background.

Bob S
 
Last edited:
  • #13
PMTs are terrible at measuring absolute light levels, particularly high gain tubes. Truly horrible. First, you are limited by the statistics of photoelectron yield at the first dynode, and second, the quantum efficiency of your photocathode varies enormously from tube to tube and even across the face of a single tube. Calibrating on single photons and scaling up introduces its own uncertainties - you have to worry about dark current and thermionic emission (i.e. current from non-light) and emission off the second dynode (i.e. reduced gain).

The only time people use them is when they have no choice. If you go that route, you will be spending 95% of your time understanding your instrumentation and 5% of your time doing your measurement.

One question you should have been asked - "How much light are you measuring?" The other is "How accurately do you need to measure it?"
 

FAQ: Can a Photomultiplier be Calibrated for Extremely Low Light Levels?

What is a photomultiplier?

A photomultiplier is a highly sensitive electronic device used to measure light intensity. It consists of a photocathode, an electron multiplier, and an anode. When light strikes the photocathode, it releases electrons which are then amplified and collected at the anode, producing an electrical signal proportional to the amount of light received.

Why is calibration necessary for a photomultiplier?

Calibration is necessary for a photomultiplier to ensure that it accurately measures light intensity. Over time, the sensitivity of a photomultiplier can change due to factors such as aging of the photocathode or changes in the electrical components. Calibration helps to correct for these changes and maintain the accuracy of the measurements.

How is a photomultiplier calibrated?

A photomultiplier is typically calibrated by exposing it to a known light source and measuring the resulting electrical signal. This can be done using a standard light source, such as a calibrated lamp, or by comparing the photomultiplier's response to that of another calibrated photomultiplier.

How often should a photomultiplier be calibrated?

The frequency of calibration for a photomultiplier depends on several factors, such as the stability of the light source being measured and the manufacturer's recommendations. In general, it is recommended to calibrate a photomultiplier at least once a year to ensure accurate measurements.

What are the potential sources of error in photomultiplier calibration?

The most common sources of error in photomultiplier calibration include variations in the light source, changes in temperature, and aging of the photomultiplier itself. Other factors, such as electrical noise and improper handling of the photomultiplier, can also contribute to errors in calibration.

Back
Top