In summary, dose calculation for an oscilloscope involves determining the amount of radiation exposure measured by the device. This process includes understanding the oscilloscope's specifications, calibrating the device, and applying appropriate formulas to convert the measured signals into dose values. Accurate dose calculation is essential for ensuring safety and efficacy in fields such as medical imaging and radiation therapy.
  • #1
Salman Khan
33
2
If pulse source output is recorded by a detector on oscilloscope, is it possible to calculate dose from this output voltage pulse?
 
Engineering news on Phys.org
  • #2
That depends on the detector and any amplifier. For an ion chamber it would be the total charge transferred, which would be the area under the curve on an oscilloscope trace under some circumstances, or the height of the pulse in others.

A Geiger counter tube would not work. Some other forms of detector might work, but interpreting the result might be challenging.
 
  • #3
Alex A said:
That depends on the detector and any amplifier. For an ion chamber it would be the total charge transferred, which would be the area under the curve on an oscilloscope trace under some circumstances, or the height of the pulse in others.

A Geiger counter tube would not work. Some other forms of detector might work, but interpreting the result might be challenging.
In case of PMT or other solid state detector?
 
  • #4
A commercial PMT/Scintillator would probably fail to work. Counters are usually designed with high value resistor chains to save power and with small value capacitors close to the anode. These saturate easily. Spectroscopy grade combinations would probably damage themselves at fairly low exposures.

If you are measuring the dose within the beam I would expect a strong pulse of light from the scintillator. Much more than would be a good fit for a PMT.

A reverse biased diode would probably give a signal, but at these power levels there may be a lot of nonlinear effects. A scintillator connected to a photodiode by a light pipe or fiber optic cable might work well.

I'm bothered that you seem to be trying to reinvent the wheel for instrumentation for a very dangerous device.
 
  • Like
Likes Salman Khan
  • #5
Dear Alex, very thanks for sharing your experience.
 
  • #6
Why use an oscilloscope? An ion chamber and a DVM works and is standard for dose measurement of x/gamma radiation.
 
  • Like
Likes Salman Khan
  • #7
So far as I know ion chamber is gaseous detector having response tim in micro second, how is it possible to measure a pulse dose of nano second pulse using ion chamber ??
 
  • #8
You didn't say nanoseconds. It works for microseconds for sure. The problem with nsec. pulses is ion recombination. A solid-state detector can be used, I think, instead of an IC, but can be calibrated with a longer pulse width beam relative to an IC.

Just found this for ICs https://www.drct.com/Ion_Chambers/TBM-IC-Pulse-X.html
 
  • Like
Likes Salman Khan

FAQ: Dose calculation for oscilloscope

What is dose calculation in the context of an oscilloscope?

Dose calculation in the context of an oscilloscope typically refers to measuring the amount of electrical signal energy or power delivered over a period of time. This can be important in various applications such as medical devices, radiation measurement, and signal integrity testing.

How do I measure the dose using an oscilloscope?

To measure the dose using an oscilloscope, you need to capture the signal waveform over the desired time period. You can then use the oscilloscope's measurement functions to calculate the area under the curve (integral) of the signal, which gives you the total energy or dose delivered.

What settings should I use on my oscilloscope for accurate dose calculation?

For accurate dose calculation, set your oscilloscope to a suitable time base and vertical sensitivity to clearly capture the entire signal waveform. Use averaging or high-resolution mode if available to reduce noise. Ensure that the probe and oscilloscope bandwidth are adequate for the signal frequency.

Can I automate dose calculation on my oscilloscope?

Many modern oscilloscopes offer built-in functions or scripts that can automate dose calculation. You can use these features to automatically integrate the signal waveform and display the dose. Additionally, some oscilloscopes support custom scripts using languages like Python or MATLAB for more complex calculations.

What are common sources of error in dose calculation using an oscilloscope?

Common sources of error include incorrect probe calibration, inadequate sampling rate, noise in the signal, and improper oscilloscope settings. Ensuring proper setup, calibration, and using noise reduction techniques can help minimize these errors.

Similar threads

Replies
1
Views
1K
Replies
9
Views
2K
Replies
4
Views
679
Replies
4
Views
1K
Replies
1
Views
1K
Replies
25
Views
4K
Replies
2
Views
2K
Replies
1
Views
1K
Back
Top