- #1
jety89
- 32
- 0
Hi, I'm working on fixing a preamplifier connected to a semiconductor detector. The high voltage bias supply is working ok, and the preamplifier preamplifies. I've confirmed this with a digital oscilloscope and a square wave generator under many different regimes. The preamplifier components are are all in place, and working, as shown here
http://muralev.narod.ru/archiv/docs/Dosimetry/PUG_schematics.JPG
X2 is the High voltage bias supply and X1 is the exit to the detector.
The proper functioning of the detector requires that " an applied voltage of typically hundreds or thousands of volts must be imposed across the active volume" (Knoll, Radiation detection and measurement).
All I can see is a flat line with very little noise, and the "behaviour" of the electronics, is... calm. I even attached the crocodile clips from the square wave generator to the bare wire at the input end of the preamplifier (after the input condensator of course), to see that it amplified in actual operating conditions, and it worked!
All I know is that the high voltage doesn't actually reach the detector, because I've measured it at several points. What I could not measure is the resistance of the resistors, since they are just too high for my multimeter, but, from what I can infer, there are three resistors between the bias supply and the detector : 100M 220M and 2x1.0G+-20%
Now, the voltages I measured went like this:
-1004V
100M
-143V
220M
-35.5V
2.0G
-5.204V
In short: the voltage actually reaching the GeLi detector is -5.2 Volts, and NOT -1000 Volts.
I am puzzled: what should I do?
I am afraid to damage the electronics, and there MUST be a reason why these very large resistors are there. My understanding is, they're there to limit the current, and they also function as filters, together with the condensators, but still: why is the voltage getting to the detector is soo low? What am I not getting here?
Additional links
http://muralev.narod.ru/archiv/docs/Dosimetry/PUG_raspolojenie.JPG
http://muralev.narod.ru/archiv/docs/Dosimetry/PUG.pdf (russian)
http://muralev.narod.ru/archiv/docs/Dosimetry/PUG_schematics.JPG
X2 is the High voltage bias supply and X1 is the exit to the detector.
The proper functioning of the detector requires that " an applied voltage of typically hundreds or thousands of volts must be imposed across the active volume" (Knoll, Radiation detection and measurement).
All I can see is a flat line with very little noise, and the "behaviour" of the electronics, is... calm. I even attached the crocodile clips from the square wave generator to the bare wire at the input end of the preamplifier (after the input condensator of course), to see that it amplified in actual operating conditions, and it worked!
All I know is that the high voltage doesn't actually reach the detector, because I've measured it at several points. What I could not measure is the resistance of the resistors, since they are just too high for my multimeter, but, from what I can infer, there are three resistors between the bias supply and the detector : 100M 220M and 2x1.0G+-20%
Now, the voltages I measured went like this:
-1004V
100M
-143V
220M
-35.5V
2.0G
-5.204V
In short: the voltage actually reaching the GeLi detector is -5.2 Volts, and NOT -1000 Volts.
I am puzzled: what should I do?
I am afraid to damage the electronics, and there MUST be a reason why these very large resistors are there. My understanding is, they're there to limit the current, and they also function as filters, together with the condensators, but still: why is the voltage getting to the detector is soo low? What am I not getting here?
Additional links
http://muralev.narod.ru/archiv/docs/Dosimetry/PUG_raspolojenie.JPG
http://muralev.narod.ru/archiv/docs/Dosimetry/PUG.pdf (russian)