I'm setting up an EPR (Electron Paramagnetic Resonance) instrument and encountering an issue with the lock-in amplifier. I’d like to share the details of the problem and receive any help or advice.
First, I’ll briefly explain the setup. (See Figure 1) On the left side of Figure 1, you can see the data measured without using the lock-in amplifier. When this data is input into the lock-in amplifier, it produces a differentiated form of data, as shown on the right side of Figure 1 in the X channel In a lock-in amplifier, if the phase is correctly adjusted, 100% of the signal should ideally appear in either the X or Y channel. I'm curious why there is still a signal remaining in the Y channel despite these adjustments.. (For a more detailed image, please refer to Figure 2.)
Ref: Zurich Instrument Hompage
This is a very ideal dataset; however, there are slight imperfections, as highlighted by the red circles.
I suspected this issue might be related to the phase-lock of the lock-in amplifier, so I changed the settings to R and Theta for the experiment. The resulting data is shown in Figure 3. In my view, Theta (Phase) should be relatively stable, with perhaps a bit of noise, but the experimental results show significant instability after the first peak. I’m wondering if this is normal data behavior or if it indicates a problem.
Another potential issue I’m considering is that our lock-in amplifier(SRS, Sr830) has a bandwidth of 100 kHz. However, the input to the lock-in amplifier consists of a 100 kHz modulated signal combined with a 100 MHz signal. Can the lock-in amplifier successfully lock onto the 100 kHz modulated signal in this case?
I thought I was familiar with lock-in amplifiers, but I realize there’s still much I don’t know. Even a small piece of advice would be a big help.
Thank you very much for reading
First, I’ll briefly explain the setup. (See Figure 1) On the left side of Figure 1, you can see the data measured without using the lock-in amplifier. When this data is input into the lock-in amplifier, it produces a differentiated form of data, as shown on the right side of Figure 1 in the X channel In a lock-in amplifier, if the phase is correctly adjusted, 100% of the signal should ideally appear in either the X or Y channel. I'm curious why there is still a signal remaining in the Y channel despite these adjustments.. (For a more detailed image, please refer to Figure 2.)
Ref: Zurich Instrument Hompage
This is a very ideal dataset; however, there are slight imperfections, as highlighted by the red circles.
I suspected this issue might be related to the phase-lock of the lock-in amplifier, so I changed the settings to R and Theta for the experiment. The resulting data is shown in Figure 3. In my view, Theta (Phase) should be relatively stable, with perhaps a bit of noise, but the experimental results show significant instability after the first peak. I’m wondering if this is normal data behavior or if it indicates a problem.
Another potential issue I’m considering is that our lock-in amplifier(SRS, Sr830) has a bandwidth of 100 kHz. However, the input to the lock-in amplifier consists of a 100 kHz modulated signal combined with a 100 MHz signal. Can the lock-in amplifier successfully lock onto the 100 kHz modulated signal in this case?
I thought I was familiar with lock-in amplifiers, but I realize there’s still much I don’t know. Even a small piece of advice would be a big help.
Thank you very much for reading