IR transmission through a very thin liquid-filled cell

In summary, it is possible to measure the transmittance of IR radiation through a cell filled with a liquid sample with thickness in the range of about 10 microns and accurately adjustable. The sample would probably have to be contained in a space between some kind of IR-transparent plates, with plate distance accurately adjustable.
  • #1
hilbert2
Science Advisor
Insights Author
Gold Member
1,598
605
Hi,

does anyone know if it's possible, with a Nicolet is50 FTIR spectrometer or similar, to measure the transmittance of IR radiation through a cell filled with a liquid sample with thickness in the range of about 10 microns and accurately adjustable? The reason for doing this would be because some substances may absorb so effectively at certain wavelengths that the transmittance would be practically zero for a larger sample thickness and one couldn't calculate the absorption coefficient for the whole wavelength range from the Lambert-Beer law.

The sample would probably have to be contained in a space between some kind of IR-transparent plates, with plate distance accurately adjustable.

Thanks,
Hilbert2
 
Chemistry news on Phys.org
  • #2
Can't you dilute the sample with an inert solvent?
 
  • Like
Likes hilbert2
  • #3
Good idea, I'll find out if that's possible. However, even an inert solvent will disturb hydrogen bonding in e.g. alcohol sample and affect the spectrum...
 
  • #4
An FTIR spectrometer or even a diffraction grating spectrometer should be able to measure the transmission spectrum.
 
  • #5
Because 10 µm is of the order of infrared wavelengths (10 µm = 1000 cm-1), your spectrum may be affected by interference fringes.
Here is a spectrum of a polyethylene film ca. 30 µ thick. At 10 µm the fringe spacing would be 3 x as great. The intensity would also depend on the refractive index mismatch at the interface.
 

Attachments

  • PE thin trans.pdf
    2 MB · Views: 334
  • Like
Likes hilbert2 and Charles Link
  • #6
Thanks, I saw those interference fringes being mentioned in some publication. The first information about this kind of measurements seems to have been produced in the 1960s: http://www.nrcresearchpress.com/doi/pdf/10.1139/v69-572 .

One approach may be to produce a liquid film with a thickness that's somewhere in the acceptable micron range, and then just measure that thickness precisely enough so that Beer-Lambert law can be used to find the absorption coefficient from the IR transmittance data.

The reason for doing these measurements is because I have to form some kind of law of how deeply blackbody-like radiation can go under a liquid surface. The radiation is produced in the gas phase above the liquid and I'm interested in something like how many millimeters of liquid are enough to absorb 90% of the total radiative energy flux - this of course depends on the blackbody temperature and the absorption spectrum of the liquid.
 
  • #7
Specific liquid or any liquid?
 
  • #9
If you are not interested in modeling of a particular substance, differences between measurements of a pure liquid and a solution can be negligible - that is, they can give values good enough to construct a model.

Then, Beer-Lambert's law is not linear for concentrated solutions, but I think you should be able to find reasonably good models that could help predict real values (at least I remember being told something like that). Or - if not - perhaps you can initially choose a rather "transparent" liquid for getting some reasonable data (transparent as in "with absorption low enough to allow measurements of more thick layers").

Just thinking aloud.
 
  • #10
Yeah, it's probably possible to find some relatively simple approximation that works well enough for complex hydrocarbon mixtures (like gasoline or motor oil) in most cases. A more problematic case would be a calculation for ethanol or some other alcohol diluted with water, as there the hydrogen bonding between the species will be important.
 
  • #11
hilbert2 said:
A more problematic case would be a calculation for ethanol or some other alcohol diluted with water, as there the hydrogen bonding between the species will be important.

Probably, but finding spectra of pure ethanol is not that difficult. I can be missing something, won't be for the first time, but I am not convinced you really need to use thin layers for recording them.

http://sdbs.db.aist.go.jp/sdbs/cgi-bin/direct_frame_disp.cgi?sdbsno=1300

(choose IR on the left)
 
  • #12
^ It's said there that the IR spectrum is of a thin ethanol film, but the actual thickness is not specified. Here's an old article where Fig. 7 shows that the absorbances of alcohols can be quite large even in 1:40 dilution with carbon tetrachloride and 0.5 mm path length: http://nvlpubs.nist.gov/nistpubs/jres/048/jresv48n4p281_A1b.pdf .
 

FAQ: IR transmission through a very thin liquid-filled cell

How does IR transmission through a very thin liquid-filled cell work?

IR transmission through a very thin liquid-filled cell works by passing an infrared light beam through a thin layer of liquid. The liquid acts as a medium through which the IR light can travel, and the light is absorbed, scattered, or transmitted depending on the properties of the liquid and the wavelength of the light.

What factors affect the transmission of IR light through a thin liquid-filled cell?

The transmission of IR light through a thin liquid-filled cell is affected by the composition and thickness of the liquid, as well as the wavelength and intensity of the light. Other factors such as temperature, pressure, and impurities in the liquid can also impact the transmission.

How is the transmission of IR light through a thin liquid-filled cell measured?

The transmission of IR light through a thin liquid-filled cell is typically measured using a spectrophotometer, which measures the amount of light that passes through the liquid. The results are often presented in a graph, showing the absorbance or transmittance of the light at different wavelengths.

What are the applications of IR transmission through a very thin liquid-filled cell?

IR transmission through a very thin liquid-filled cell has various applications in fields such as chemistry, biology, and environmental science. It is commonly used for qualitative and quantitative analysis of liquid samples, as well as for process monitoring and control in industries such as pharmaceuticals and food processing.

Can IR transmission through a very thin liquid-filled cell be used for non-destructive testing?

Yes, IR transmission through a very thin liquid-filled cell can be used for non-destructive testing. It allows for the analysis of liquid samples without altering or damaging them, making it a valuable tool in various industries for quality control and research purposes.