Does anyone have experience with radiation doses and measurement units in radiology?
I was looking at some DICOM metadata...
X-ray image of sinuses that include forehead sinus using Philips DigitalDiagnost C50
"Image And Fluoroscopy Area Dose Product" stood out to me since it is measured in dGy - deciGray
When converting miliGray (mGy) to miliSievert, that would be 163.7 mSv?
VS.
Cranium/Head CT radiation dose. Using GE Revolution Maxima
This CT study is around 1.680-1.9 mSv according to this DLP calculator, in some other places on the internet, it is mentioned head CT is typically rounded to 2 mSv, which is 8 months of natural background radiation.
DICOM from CT shows much lower X-ray tube current (only 10 mA compared to 581 mA in traditional xray), but longer exposure time in msec (there are various parameters, several exposure times) and therefore larger 'exposure', which is just a pure calculation in mAs.
It seems that according to some DICOM info, "Image And Fluoroscopy Area Dose Product" is automatically measured with a specific capturing device built in/behind the surface where you are lying or standing in front of..
But could it really be that you receive a higher X-ray dose from a traditional head or sinus X-ray than from a full head CT?
I was looking at some DICOM metadata...
X-ray image of sinuses that include forehead sinus using Philips DigitalDiagnost C50
(0018,0015) Body Part Examined SKULL
(0018,0060) KVP 77
(0008,1030) Study Description Sinuses
(0008,103E) Series Description Open Mouth PA
(0018,1150) Exposure Time 12 ms
(0018,1151) X Ray Tube Current 581 mA
(0018,1152) Exposure 7 mAs
(0018,1153) Exposure in µAs 6680
(0018,115E) Image And Fluoroscopy Area Dose Product 1.673
"Image And Fluoroscopy Area Dose Product" stood out to me since it is measured in dGy - deciGray
Image and Fluoroscopy Area Dose Product Attribute - X-Ray dose, measured in dGy*cm*cm, to which the patient was exposed for the acquisition of this image plus any non-digitally recorded fluoroscopy that may have been performed to prepare for the acquisition of this image.
1 dGy = 100 mGy,
so 1.673 (dGy × cm2) = 167.3 mGy
When converting miliGray (mGy) to miliSievert, that would be 163.7 mSv?
1 milligray [mGy] = 0.001 sievert [Sv]
167.3 milligray = 0.1673 sievert or 163.7 mSv
VS.
Cranium/Head CT radiation dose. Using GE Revolution Maxima
Dose Report
Series | Type | Scan Range (mm) | CTDIvol (mGy) | DLP (mGy-cm) | Phantom cm |
---|---|---|---|---|---|
1 | Scout | S160.000-l100.000 | 0.07 | 1.85 | Body 32 |
2 | Scout | S160.000-I100.000 | 0.07 | 1.85 | Body 32 |
3 | Axial | l21.000-S56.500 | 45.75 | 366.03 | Head 16 |
4 | Axial | S58.000-S153.000 | 43.03 | 430.27 | Head 16 |
Total Exam DLP: | 800.01 |
This CT study is around 1.680-1.9 mSv according to this DLP calculator, in some other places on the internet, it is mentioned head CT is typically rounded to 2 mSv, which is 8 months of natural background radiation.
DICOM from CT shows much lower X-ray tube current (only 10 mA compared to 581 mA in traditional xray), but longer exposure time in msec (there are various parameters, several exposure times) and therefore larger 'exposure', which is just a pure calculation in mAs.
(0018,1150) Exposure Time 8000
(0018,1151) X Ray Tube Current 10
(0018,1152) Exposure 26
...
(0018,0015) Body Part Examined HEAD
(0018,0060) KVP 120
(0018,1150) Exposure Time 8000
(0018,8151) X Ray Tube Current in µA 127770.74
...
(0018,0015) Body Part Examined HEAD
(0018,0060) KVP 120
(0018,1150) Exposure Time 10000
(0018,8151) X Ray Tube Current in µA 120156.83
It seems that according to some DICOM info, "Image And Fluoroscopy Area Dose Product" is automatically measured with a specific capturing device built in/behind the surface where you are lying or standing in front of..
But could it really be that you receive a higher X-ray dose from a traditional head or sinus X-ray than from a full head CT?