A Photon Bouncing during APD Dead Time in Autocorrelation Measurement

  • A
  • Thread starter Thread starter Theo2907
  • Start date Start date
AI Thread Summary
The discussion centers on challenges faced in an autocorrelation g(2) measurement setup using Avalanche Photodiodes (APDs) and a multimode fiber system. The researcher is observing unexpected peaks in dark counts, specifically at 20 ns and 80 ns, which are believed to be related to the dead time of the APDs and potential reflections within the fiber couplings. There is a hypothesis that photons may be bouncing back through the fiber during the APD dead time, causing interference. Suggestions to switch to single-mode fibers have been made, but the researcher is concerned about reduced collection efficiency for future measurements. Insights and recommendations for optimizing APD setups in autocorrelation measurements are being sought to address these issues.
Theo2907
Messages
1
Reaction score
4
TL;DR Summary
Encountering strange results in dark counts of APD. Suspect reflections during APD dead time causing interference. Seeking quick insights to mitigate this issue. Any advice is appreciated.
Hey fellow researchers,

I'm currently working on an autocorrelation g(2) measurement setup using two Avalanche Photodiodes (APDs) connected through a multimode GRIN 50:50 beam splitter. Photons are delivered to the system via a multimode fiber (a PC-PC coupling to the beam splitter). However, I've been encountering a challenge and I believe it is related to the dead time of the APDs.
Below you can find a picture of the setup:
1702372340504.png


Before doing a complete g(2) measurement I wanted to check if my dark counts are incoherent (which should be the case). Unfortunately, I found that they are not incoherent, I did a measurement measuring the time difference between a measurement at APD 1 and APD 2 and I got peaks at around 20 ns and 80 ns.
You can see a screenshot of the graph below:

1702373825349.png


I couldn't explain it and no one in my lab has a good explanation for this phenomenon.

The peak at 20ns only happens when the beam splitter is actually coupled to the APDs, so a dark count measurement with the APDs not coupled to anything is completely incoherent (which is what we want).
The 80ns peak only happens when the multimode fiber(not present in the picture) is connected to the input fiber.

I do have a hypothesis though about this but I can't find any source online that verifies this.

I think that during the dead time of the APDs, incoming photons are being reflected, causing them to bounce back through the fiber. They are then reflected again at fiber couplings, the first one being the coupling between the input fiber and the MM fiber and there is another coupling from the MM fiber to another MM fiber.

Some people have advised me to use single-mode fibers instead and APC-APC couplings but if I use single-mode fiber my collection efficiency is not high enough and I don't get enough counts to make any measurement so that is not really an option. Later I want to measure light from single photon emitters in 4H-SiC so the collection efficiency is quite important.

I would greatly appreciate any insights, advice, or experiences you may have regarding this issue. Have you encountered similar challenges in your experiments? How did you address or mitigate the effects of this phenomenon?

Additionally, if anyone has recommendations for optimizing APD setups in autocorrelation measurements, please feel free to share. I'm open to suggestions and eager to learn from your expertise.

Thanks in advance for your help!
 
  • Like
Likes PeroK and berkeman
Physics news on Phys.org
Theo2907 said:
TL;DR Summary: Encountering strange results in dark counts of APD. Suspect reflections during APD dead time causing interference. Seeking quick insights to mitigate this issue. Any advice is appreciated.

I think that during the dead time of the APDs, incoming photons are being reflected, causing them to bounce back through the fiber.
I see there have been no responses yet so here's one: The time between the first small peak and the main peak is about 60ns. That represents a large path delay (20m??) and your image suggests you should expect less than 1m path lengths for any reflections. Could it be an instrumentation problem?
 
Thread 'Is there a white hole inside every black hole?'
This is what I am thinking. How much feasible is it? There is a white hole inside every black hole The white hole spits mass/energy out continuously The mass/energy that is spit out of a white hole drops back into it eventually. This is because of extreme space time curvature around the white hole Ironically this extreme space time curvature of the space around a white hole is caused by the huge mass/energy packed in the white hole Because of continuously spitting mass/energy which keeps...