In summary, using a katharometer to measure the amount of oxygen in a mixture of nitrogen and hydrogen requires a constant proportion of hydrogen in order to limit errors to + 0.1%. The thermal conductivity is proportional to the composition, so the k value of the nitrogen / hydrogen mix with oxygen (5%) would be 1.02445. To find the percentage change of hydrogen that would give a 0.1% change in oxygen, the k value of the nitrogen / hydrogen mix with oxygen (5.1%) would be 1.024479.
cjm181
69
It is proposed to use a katharometer to measure the amount (about 5%) of oxygen in nitrogen, in the presence of a small amount (0.5%) hydrogen. How constant would the proportion of hydrogen have to be in order to limit errors in measurement of % oxygen to + 0.1%?The thermal conductivities are:
nitrogen 0.993
oxygen 1.052
hydrogen 6.993
So I understand from the notes that the thermal conductivity is proportional to the composition. So, the thermal conductivity of nitrogen (99.995%) and hydrogen (0.005%) would be:
6.993 - 0.993 =6
6 * 0.005 = 0.03
0.993 + 0.03 = 1.023
So nitrogen (99.995%) mixed with hydrogen (0.005%) would have a k value of 1.023
So, to find the K value of the nitrogen / hydrogen mix with oxygen (5%):
1.052 - 1.023 = 0.029
0.029 & 0.05 = 0.00145
0.00145 + 1.023 = 1.02445
So the value of nitrogen / hydrogen (95%) and oxygen (5%) is 1.02445
So, to find the percentage of hydrogen change that would give a 0.1% change in oxygen? that's where I am stuck!
If the oxygen content changed by 0.1%, then the k value would be:
To find the K value of the nitrogen / hydrogen mix with oxygen (5.1%):
0.029 & 0.051 = 0.001479
0.001479 + 1.023 = 1.024479
So I need to find the percentage change of hydrogen that would give a total k value of 1.024479?
Related to Solve Katharometer Question: Limit Errors to +0.1% Oxygen
1. What is a Katharometer?
A Katharometer is a scientific instrument used to measure the concentration of a gas, specifically oxygen, in a gas mixture.
2. How does a Katharometer work?
A Katharometer uses a heated filament to ionize oxygen molecules in the gas mixture. The change in resistance of the filament is then measured and used to calculate the concentration of oxygen.
3. How do you solve for Katharometer question?
To solve a Katharometer question, you will need to know the limit error of the instrument (+0.1% in this case) and the measured concentration of oxygen. Then, you can use the formula: Limit Error = (measured concentration - true concentration) / true concentration * 100% to solve for the true concentration of oxygen.
4. What factors can affect the accuracy of a Katharometer?
The accuracy of a Katharometer can be affected by factors such as temperature changes, flow rate of the gas mixture, and impurities in the gas mixture.
5. How can you minimize errors when using a Katharometer?
To minimize errors, it is important to calibrate the Katharometer regularly and ensure that it is properly maintained. Additionally, controlling and monitoring the temperature and flow rate of the gas mixture can also help reduce errors.
10K