SR830 Lock In Amplifier Not Reading Reference Signal

In summary, the conversation discusses issues with the SR830 lock-in amplifier not picking up the reference signal and not stabilizing the output voltage in a measurement setup. Suggestions are given to reduce noise, increase signal amplitude, adjust sensitivity range, increase time constant, and turn on filters for better measurements.
  • #1
dkuljis
2
0
So here is my dilemma:

We have our SR830 in a MOKE set-up. The modulation of the laser is with a chopper at approximataly 336 Hz. The reference signal is not being picked up by the SR830, but using the same cable the reference frequency is coming up on our oscilloscope. ie, plug the cable into the oscilloscope or DMM and we get the appropriate frequency. Plug it into the SR830, and the "Unlock" light is on with a frequency of exactly 1 kHz, which my adviser says it means it's not getting a signal. The rest of the SR830 seems to be in good working order. Both CH1 and CH2 are picking up signals from our photo detector. Any thoughts or suggestions?

Sorry if the post is skim on details, I'm new to all this.

Thanks for your time,
Danny
 
Engineering news on Phys.org
  • #2
"The SR810 and SR830 Lock-In Amplifiers have differential inputs with 6 nV/√Hz input noise. The input impedance is 10 MΩ, and minimum full-scale input voltage sensitivity is 2 nV. "

WOW that's some instrument.

The reference input needs to be 400mV pp minimum. You didn't mention what the voltage is.

Has the setup ever worked?

Does the instrument work in other setups with an external reference?

Can you drive the external reference with a signal generator to test its acquisition range?

Just guessing.
 
  • #3
Thanks meBigGuy. We measured the voltage on the reference signal and it was only around 140 mV so we switched our detector and got the reference frequency where it should be. Thanks a bunch for the quick response!Sent from my iPhone using Physics Forums
 
  • #4
I am measuring voltage output from the pick-up coils of a vibrating sample magnetometer using the SR830 lock-in amplifier. The problem is that the channel 1 output voltage cannot be stabilized at all. I read the manual but that is not helpful either. What can I do to stabilize the reading ?
 
  • #5
Sakin_EEE said:
I am measuring voltage output from the pick-up coils of a vibrating sample magnetometer using the SR830 lock-in amplifier. The problem is that the channel 1 output voltage cannot be stabilized at all. I read the manual but that is not helpful either. What can I do to stabilize the reading ?

*Reduce the amount of noise in your measurement setup. Lock-ins can "cut through" quite a lot of noise, but there are always limits and this should always be the first step
*Increase the amplitude of the input signal
*Make sure you have the SR830 is on the optimal sensitivity range
*Increase the time constant
*Turn on the filters on the input.

Generally speaking, there will always be some trade-off between measurement speed and sensitivity; if your noise is mainly white you can detect some very small signals with a lock-in if you increase the time constant.
 
  • Like
Likes Sakin_EEE and berkeman

FAQ: SR830 Lock In Amplifier Not Reading Reference Signal

1. Why is my SR830 Lock In Amplifier not reading the reference signal?

There could be several reasons for this issue. First, make sure that the reference signal is connected to the reference input of the amplifier. Also, check the cables and connections to ensure they are secure and not damaged. If the issue persists, try adjusting the input sensitivity and time constant settings on the amplifier. If none of these solutions work, there may be a problem with the amplifier itself and it may need to be repaired or replaced.

2. Can I use a different reference signal with my SR830 Lock In Amplifier?

Yes, the SR830 Lock In Amplifier has a wide range of input frequencies that it can accept as a reference signal. However, it is important to make sure that the input frequency falls within the specified range for accurate readings.

3. How do I calibrate the reference signal on my SR830 Lock In Amplifier?

To calibrate the reference signal, you will need to use an external frequency standard or a stable AC voltage source. Connect the reference signal to the amplifier and go to the "Setup" menu. Select "Reference" and then "External" to choose the external source. Press "Auto Gain" to adjust the input sensitivity, and then press "Auto Phase" to adjust the phase. Your reference signal should now be calibrated.

4. What does it mean if my SR830 Lock In Amplifier is showing a low reference signal quality?

This could indicate that the reference signal is weak or noisy. Check the cables and connections to ensure they are secure and not damaged. You may also need to adjust the input sensitivity and time constant settings on the amplifier to improve the signal quality.

5. Can I use the internal oscillator as a reference signal on my SR830 Lock In Amplifier?

Yes, the SR830 Lock In Amplifier has an internal oscillator that can be used as a reference signal. However, it is important to note that this oscillator is not as stable as an external frequency standard, so it may not be suitable for all applications.

Similar threads

Replies
5
Views
1K
Replies
10
Views
3K
Replies
3
Views
2K
Replies
4
Views
4K
Replies
1
Views
5K
Replies
1
Views
2K
Replies
12
Views
11K
Back
Top