The multimeter is not displaying mv of strain gauge in circuit

In summary: Thank you very much, hx711 supplying voltage is 5v, i shoud use dc power supply and give 1.5v to both hx711 and wheatstone brdige
  • #1
Micheal_Leo
79
4
i have made the wheatstone bridge( 3 resistors and 1 strain gauge) circuit but multimiter is not displaying anything , please need guide
 

Attachments

  • h1.jpg
    h1.jpg
    43.5 KB · Views: 68
  • h2.jpg
    h2.jpg
    35.8 KB · Views: 82
  • h3.jpg
    h3.jpg
    42.5 KB · Views: 60
Physics news on Phys.org
  • #2
Micheal_Leo said:
... but multimiter is not displaying anything ...
You have an overload on the mV scale. Switch to DC volts to see what voltage you have.

Please provide the circuit you are using with all resistor values.
What is the resistance of the strain gauge?
 
  • Like
Likes Micheal_Leo
  • #3
Baluncore said:
You have an overload on the mV scale. Switch to DC volts to see what voltage you have.
Please provide the circuit you are using with all resistor values.
What is the resistance of the strain gauge?

please find the attached schematic diagram
 

Attachments

  • wx_camera_1690883390131.jpg
    wx_camera_1690883390131.jpg
    12.8 KB · Views: 75
Last edited by a moderator:
  • #4
The resistor ratios must match.
Is that a pair of 220R = 220 ohm resistors in the top row, making half the supply ?

Below left is another 220R resistor, but it should match the 120 ohm strain gauge, so it also makes about half the supply. Then mV range should work.

Replace the bottom left 220R with a 120R resistor.
 
  • #5
Baluncore said:
The resistor ratios must match.
Is that a pair of 220R = 220 ohm resistors in the top row, making half the supply ?

Below left is another 220R resistor, but it should match the 120 ohm strain gauge, so it also makes about half the supply. Then mV range should work.

Replace the bottom left 220R with a 120R resistor.
yes half supply,i do not have 120 ohm resistors,.can i use 100+ 22 ohm resistors in series?
 
  • #6
Micheal_Leo said:
please find the attached schematic diagram
Why are you using such a high bias voltage? Maybe switch to a single 1.5V AA battery to avoid wasting so much power...
 
  • #7
berkeman said:
Why are you using such a high bias voltage? Maybe switch to a single 1.5V AA battery to avoid wasting so much power...

actually i have limited resources so have to do in limit resource
 
Last edited by a moderator:
  • #8
Micheal_Leo said:
yes half supply,i do not have 120 ohm resistors,.can i use 100+ 22 ohm resistors in series?
Yes.
 
  • #9
Baluncore said:
Yes.
THE Attached results are mv and Voltage , please see this , i will connect it to hx711, these results before force not applied , please guide that should i consider mv or voltage to transfer it to arduino to study stress in that part.the PLA printing material not follow hookes law i have applied force on part for testing so little bending is present in part
 

Attachments

  • Weixin Image_20230802140610.jpg
    Weixin Image_20230802140610.jpg
    41.4 KB · Views: 66
  • Weixin Image_20230802140602.jpg
    Weixin Image_20230802140602.jpg
    42.5 KB · Views: 79
  • Weixin Image_20230802141542.jpg
    Weixin Image_20230802141542.jpg
    26.9 KB · Views: 80
  • #10
I do not know where you are going with this.
Do you see any change in the mV reading when you flex the strain gauge?

If you see only a small change, you may need to adjust the 22R resistor, to better balance the bridge, to get closer to a zero difference voltage output. Then you can use more gain in the hx711 without the amplified DC offset hitting the voltage limits.

The differential voltage output that now goes to your mV meter, will go to the hx711 differential inputs, INA+, INA-. The bridge drive voltage will probably come from the hx711, AVDD and AGND.
 
  • #11
Baluncore said:
I do not know where you are going with this.
Do you see any change in the mV reading when you flex the strain gauge?

If you see only a small change, you may need to adjust the 22R resistor, to better balance the bridge, to get closer to a zero difference voltage output. Then you can use more gain in the hx711 without the amplified DC offset hitting the voltage limits.

The differential voltage output that now goes to your mV meter, will go to the hx711 differential inputs, INA+, INA-. The bridge drive voltage will probably come from the hx711, AVDD and AGND.
i see change in mv but also my finger burn accidently comes near to strain gauge , strain gauge get hot during this phenomena mv voltage out of control like it kept increasing ,

should i give 5V , right now i am giving 9V
 
  • #12
Micheal_Leo said:
... i see change in mv but also my finger burn accidently comes near to strain gauge ...
Too much voltage to the bridge circuit.
Too much bridge current.
Too much heating of the strain gauge resistance.
The gauge resistance is temperature dependent, so balance changes.
Use less voltage, use AVDD and AGND from the hx711.
 
  • Like
Likes Micheal_Leo
  • #13
Baluncore said:
Too much voltage to the bridge circuit.
Too much bridge current.
Too much heating of the strain gauge resistance.
The gauge resistance is temperature dependent, so balance changes.
Use less voltage, use AVDD and AGND from the hx711.
thank you very much, hx711 supplying voltage is 5v, i shoud use dc power supply and give 1.5v to both hx711 and wheatstone brdige
 
  • #14
The hx711 can run on between 3 and 5 V. If you use the same voltage as your controller, the digital signal voltages will all match without problems.

The hx711 has a regulator that provides the voltage drive to the bridge. The voltage is set by the ratio of R1 and R2 on the hx711 reference PCB design in the data sheet. Use what it has now, or select resistors that set the bridge voltage to somewhere between 1.2 and 2.5 volts.
 
  • Like
Likes Micheal_Leo
  • #15
Baluncore said:
The hx711 can run on between 3 and 5 V. If you use the same voltage as your controller, the digital signal voltages will all match without problems.

The hx711 has a regulator that provides the voltage drive to the bridge. The voltage is set by the ratio of R1 and R2 on the hx711 reference PCB design in the data sheet. Use what it has now, or select resistors that set the bridge voltage to somewhere between 1.2 and 2.5 volts.
i have voltage regulator so have intention to supply 3v to both bridge and hx711
 
  • #16
Micheal_Leo said:
i have voltage regulator so have intention to supply 3v to both bridge and hx711
If you increase the bridge drive voltage by 1%, you get a 1% increase in the difference voltage from the bridge, for the same strain. That is a problem with accurate 24 bit measurements.

The A-D converter is ratiometric. It uses the bridge drive voltage as a full-scale reference, and so automatically corrects for variations. That is why it has the voltage regulator for the bridge drive on the same chip as the A-D converter. If you do not use that on-chip regulator, you will get scale errors in the strain readings. Read the data sheet, and follow the advice.
 
  • Like
Likes berkeman
  • #17
Baluncore said:
If you increase the bridge drive voltage by 1%, you get a 1% increase in the difference voltage from the bridge, for the same strain. That is a problem with accurate 24 bit measurements.

The A-D converter is ratiometric. It uses the bridge drive voltage as a full-scale reference, and so automatically corrects for variations. That is why it has the voltage regulator for the bridge drive on the same chip as the A-D converter. If you do not use that on-chip regulator, you will get scale errors in the strain readings. Read the data sheet, and follow the advice.
please can you suggest best A-D converter
 
  • #18
Micheal_Leo said:
please can you suggest best A-D converter
You cannot afford the best.
Use the hx711, because it is available, cheap ($2), and comes mounted on a PCB.
 
  • Like
Likes Micheal_Leo
  • #19
Baluncore said:
You cannot afford the best.
Use the hx711, because it is available, cheap ($2), and comes mounted on a PCB.
i have try to do it , somehow it work as i expected , the video attached , secondly the follwoing hx711 amplifier i have , i am very much confused how to attach wheatstone bridge with this hx711 , please need your guidance

video about strain gauge and wheatstone bridge

i have try to draw diagram of connecting , please check
 

Attachments

  • Weixin Image_20230803090826.jpg
    Weixin Image_20230803090826.jpg
    31 KB · Views: 72
  • Weixin Image_20230803090833.jpg
    Weixin Image_20230803090833.jpg
    19.9 KB · Views: 66
  • gt.jpg
    gt.jpg
    16.7 KB · Views: 93
Last edited:
  • #20
I think, in the Weixin picture, at the analogue end of the PCB ...
A+ and A- are the bridge output terminals, (your old meter connections).
GND is probably analogue ground to the bridge.
Out+ is the regulated power to the bridge.
You can check those by following the terminals to the chip pins.

Do you have a copy of the hx711 data sheet?
It shows the connections to the bridge on page 1, fig 1. Also page 6, fig 4.
https://cdn.sparkfun.com/datasheets/Sensors/ForceFlex/hx711_english.pdf
 

FAQ: The multimeter is not displaying mv of strain gauge in circuit

Why is my multimeter not displaying the millivolt (mV) reading from the strain gauge?

There could be several reasons for this issue: the strain gauge might not be properly connected, the multimeter could be set to the wrong measurement mode, or there might be an issue with the strain gauge itself. Double-check all connections and ensure the multimeter is set to measure millivolts.

How can I verify that my strain gauge is working correctly?

To verify that your strain gauge is working, you can use a known load or stress to see if the strain gauge produces a corresponding change in resistance. You can also check the resistance of the strain gauge with a multimeter to ensure it matches the expected value when not under load.

What settings should I use on my multimeter to measure the output of a strain gauge?

Set your multimeter to measure DC millivolts (mV). Ensure that the range is appropriate for the expected output of the strain gauge. If your multimeter has an auto-range feature, enable it to simplify the process.

Could the issue be related to the power supply for the strain gauge circuit?

Yes, the power supply could be a factor. Ensure that the strain gauge is receiving the correct excitation voltage, typically provided by a stable DC power source. An unstable or incorrect power supply can lead to inaccurate or no readings.

Is it possible that the multimeter itself is faulty?

It's possible that the multimeter is faulty. To check, try measuring a known voltage source with the multimeter. If the multimeter does not display the correct voltage, it may need calibration or repair. Alternatively, try using a different multimeter to see if the issue persists.

Similar threads

Back
Top