Troubleshooting Proportional Counter Problem

In summary, the conversation revolved around a problem with a proportional Gas detector that has been degrading over the past two months, causing issues with Mossbauer experiments. The discussion included suggestions to check the gas circulation, decay of the source, and potential issues with the amplifier and power supply. It was also recommended to consult a book on gas detectors for further information.
  • #1
Xaritou
6
0
Hi guys. I have a problem with my spectrum and I would like your help. I have a system with a proportional Gas detector (krypton gas), preamplifier and amplifier, Co57 source, and the last two months the spectrum has gotten worst as you can see in the attachment. Does anyone know what might be the problem? Detector? Preamplifier? Unfortunately I don't have spare parts so I can't eliminate something.
The last two pictures are the output signal from the preamplifier, and the amplifier
Co57 Spectrum.jpg
20150127_155518.jpg
20150127_155411.jpg
 
Physics news on Phys.org
  • #2
I'd look for the manufacturer's website and see if someone has posted this issue there.
 
  • #3
This obviously depends on how much you've used it, but how's the gas circulation in the counter?

To me, it looks like the FWHM of the peaks is pretty much the same, ditto the relative heights of the peaks. Were the spectra taken for the same time (accounting for a little 57Co decay)? It just looks like you've just lost a little gain? Is this spectrum actually unusable to you?

jedishrfu - that implies it has been bought off the shelf! :D
 
  • #4
e.bar.goum said:
This obviously depends on how much you've used it, but how's the gas circulation in the counter?

To me, it looks like the FWHM of the peaks is pretty much the same, ditto the relative heights of the peaks. Were the spectra taken for the same time (accounting for a little 57Co decay)? It just looks like you've just lost a little gain? Is this spectrum actually unusable to you?

jedishrfu - that implies it has been bought off the shelf! :D

Yes the spectra was taken for the same time, the difference from the decay should be very small. The detector is a "closed can" style, so there is no gas circulation. The spectrum is not unusable but its not perfect. I must use the counts from the 14keV peak, for Mossbauer experiments, and the "non 14keV" i take into account are extra noise so it creates a little problem.

The thing that annoys me, is that the peak at 6 keV remained perfect. an the other two are problematic.
 
  • #5
It seems to me that the difference due to decay in one month should be about 8%. (The halflife of 57Co is only 272 days, after all). But that's probably not the only problem. I'm assuming you've kept the geometry constant as well.

I asked about gas flow because you do expect that closed-can type proportional counters will degrade over time due to deposition.

It's hard to tell without being there, obviously, and it's hard to tell without being able to swap out components. I'd be taking away the source, and looking at what the scope is triggering on - are you picking up any RF? If you move around the amplifier/preamp/cables, does the noise change?

ETA: Another thought: How's the power supply? Has your leakage current gone substantially up?
 
Last edited:
  • #6
e.bar.goum said:
It seems to me that the difference due to decay in one month should be about 8%. (The halflife of 57Co is only 272 days, after all). But that's probably not the only problem. I'm assuming you've kept the geometry constant as well.

I asked about gas flow because you do expect that closed-can type proportional counters will degrade over time due to deposition.

It's hard to tell without being there, obviously, and it's hard to tell without being able to swap out components. I'd be taking away the source, and looking at what the scope is triggering on - are you picking up any RF? If you move around the amplifier/preamp/cables, does the noise change?

ETA: Another thought: How's the power supply? Has your leakage current gone substantially up?

I changed the cables completely, the power supply is ok. I even tried it with 1500-1800 V (it was 1700 originally). Without source, its just little random noise. Geometry is constant. I also removed the cryostat, shielding and collimator. At first. I though it was x ray fluorescence from something.
It looks to me that it actually split the main peaks in two, with the second one in little less energy. Maybe the ions can't move and be collected fast enough so they create an electric field that "reduce" the next avalanche and the measured energy from the next photon.
But I don't know the symptoms from gas degradation. If someone can suggest a book or something to read about gas detectors and problems, I would be grateful.
 
  • #7
Xaritou said:
I changed the cables completely, the power supply is ok. I even tried it with 1500-1800 V (it was 1700 originally). Without source, its just little random noise. Geometry is constant. I also removed the cryostat, shielding and collimator. At first. I though it was x ray fluorescence from something.
It looks to me that it actually split the main peaks in two, with the second one in little less energy. Maybe the ions can't move and be collected fast enough so they create an electric field that "reduce" the next avalanche and the measured energy from the next photon.
But I don't know the symptoms from gas degradation. If someone can suggest a book or something to read about gas detectors and problems, I would be grateful.

It sounds like you've done the obvious things. I highly recommend Radiation Detection and Measurement by Glenn F. Knoll as a first resource for any detector theory. There's a chapter on proportional counters. He has a comprehensive set of references at the end of every chapter too, so you can use it as a springboard for further reading.
 

FAQ: Troubleshooting Proportional Counter Problem

What is a proportional counter?

A proportional counter is a type of gas-filled radiation detector that measures the number of ionization events caused by incoming radiation. It is used in various scientific and medical applications, such as particle accelerators and X-ray machines.

What are some common problems with proportional counters?

Some common problems with proportional counters include loss of gas pressure, high voltage breakdown, and contamination of the detector window. These issues can cause the detector to malfunction and provide inaccurate readings.

How do you troubleshoot a proportional counter problem?

The first step in troubleshooting a proportional counter problem is to check the gas pressure and make sure it is at the correct level. If the pressure is low, the detector may need to be refilled. Next, check the high voltage power supply and make sure it is functioning properly. If the detector window is contaminated, it can be cleaned with an appropriate solvent.

What should I do if my proportional counter is still not working after troubleshooting?

If the problem persists after troubleshooting, it may be necessary to replace certain components, such as the gas or the detector window. It is also important to ensure that the detector is properly calibrated and that the correct settings are being used for the type of radiation being measured.

How often should a proportional counter be checked for potential problems?

It is recommended to perform routine maintenance and checks on a proportional counter at least once a year. This includes checking the gas pressure, cleaning the detector window, and calibrating the detector. If any abnormalities or malfunctions are noticed during this process, the detector should be checked more frequently and repaired as needed.