What factors affect total dose at dmax in parallel-opposed fields?

  • Thread starter hkabc
  • Start date
In summary: This also results in a lower dose to the spinal cord, making 10 MV photons preferable for treating lung volumes with off-cord parallel opposed-oblique fields. In summary, the PDD and total dose at dmax are affected by factors such as photon energy, field size, and patient thickness. Using 10 MV photons is preferable for treating lung volumes with off-cord parallel opposed-oblique fields, as it results in a lower total dose at dmax and a lower dose to the spinal cord.
  • #1
hkabc
1
0
Dear members,
I am frustrated by the Raphex exam questions. there are many I cannot solved. And the answer section is very brief for most questions.

T 10. The total dose delivered at depth dmax, from a pair of parallel-opposed fields, expressed as a percentage of the total dose at mid-plane:
A. Decreases as photon energy increases.
B. Decreases as field size increases
C. Increases as patient thickness increases.
D. Is slightly less for an SSD setup than for an SAD setup.
E. All of the above.

T 11 . When treating a lung volume with off-cord parallel opposed-oblique fields, 10 MV photons
would be preferable to 6 MV photons because:
A. 6 MV would give inadequate dose in the build-up region.
B. Lung corrections are more accurate with higher energy.
C. The total dose at dmax is less with 10 MV
D. The cord dose is less with 10 MV

ANSWERS:
T10. E Any factor that increases the PDD will decrease the total dose at dm,,x, compared with the total dose at midplane. Treating at SSD rather than SAD gives a slightly higher PDD.
T 11 . C For parallel-opposed fields, the total dose at dm,,X, as a percentage of the dose at the isocenter, is lower for higher photon energy.


Could anyone elaborate, please?
 
Physics news on Phys.org
  • #2
For T10, the answer is that all of the factors mentioned, such as increasing photon energy, field size, and patient thickness, will lead to an increase in the PDD (percentage depth dose). This increase in PDD will cause a decrease in the total dose at dmax, compared to the total dose at midplane. Treating at SSD rather than SAD will give a slightly higher PDD, which leads to a slight decrease in the total dose at dmax compared to that of an SAD setup.For T11, the answer is that for parallel-opposed fields, the total dose at dmax, as a percentage of the dose at the isocenter, is lower for higher photon energy. This means that using 10 MV photons, instead of 6 MV, would result in a lower total dose at dmax. This is because the PDD is lower for higher photon energies, leading to a decrease in the total dose at dmax compared to the total dose at the isocenter.
 

FAQ: What factors affect total dose at dmax in parallel-opposed fields?

1. What is Raphex 2005?

Raphex 2005 is a set of questions and problems used in the annual Radiological Physics Examination (RAPHEX) for medical physicists. It covers a range of topics related to medical imaging, radiation therapy, and radiation safety.

2. Who uses Raphex 2005?

Raphex 2005 is primarily used by medical physicists as a practice tool for the RAPHEX exam. However, it can also be used by students and professionals in the field of radiological physics to test their knowledge and understanding of key concepts.

3. How many questions are included in Raphex 2005?

There are a total of 100 questions in Raphex 2005, divided into 10 sections with 10 questions each. Each section focuses on a different topic, such as radiation safety, imaging modalities, and quality control.

4. Are the questions in Raphex 2005 updated regularly?

No, the questions in Raphex 2005 have not been updated since the exam was administered in 2005. However, the concepts and principles covered in the questions are still relevant and important for medical physicists to understand.

5. Can Raphex 2005 be used as a study guide for other exams?

Raphex 2005 is specifically designed for the RAPHEX exam and may not cover all topics and concepts required for other exams. It is best used as a supplement to other study materials and should not be relied upon as the sole source of preparation for other exams.

Similar threads

Back
Top