Wheatstone bridge with four active strain gauges - extension cable length compensation formula

In summary, the Wheatstone bridge configuration with four active strain gauges is utilized to accurately measure strain in various applications. The extension cable length compensation formula is critical to ensure that the measurements remain precise despite variations in cable length, which can introduce errors due to resistance changes. This formula accounts for the resistance of the cables and adjusts the readings accordingly, thereby enhancing the reliability of the strain gauge measurements.
  • #1
MarsBravo
3
0
Suppose a Wheatstone bridge with four equal active strain sensors (350 ohm each).
The output nodes are connected to a high impedance instrumentation amplifier, so no current draw from there. The bridge is placed at consiberable distance from the exitation voltage source (and the amplifier), so the length of the extension cable (with all four wires) must be accounted for in such a way that it adds to the total resistance of the bridge but not to the sensitive part (the bridge). It will reduce the sensivity of the bridge, but in a small way.
Can the bridge and the extension cable resistors considered as three resistors in series (the bridge resistance measured from the exitation source is still 350 ohm!), or has one take into account the split of resistors within the bridge (dividing the exitation current) as the extension cable resistors 'serves' both branches?
In the attachment, the difference in the formula is in red. CF is the calibration factor, CF' is the adjusted CF. Which formula is right?
 

Attachments

  • WBCC.pdf
    177.9 KB · Views: 85
Engineering news on Phys.org
  • #2
The resistance of the wires, from the bridge to the amplifier, play no part in the sensitivity. The voltage across the bridge is reduced by the lead resistance, which reduces the excitation current.

The 4 resistor bridge looks like one resistor to the excitation current, Ie.
The sensitivity of the bridge is proportional to Ie. Which is reduced by the series voltage drop in the wires.

Without wires, Ie = Ve / Rb;
With two wires of resistance, Rw;
Ie = Ve / ( Rb + Rw + Rw );
Ie and the sensitivity will be reduced to; Rb / ( Rb + Rw + Rw );
350 / ( 350 + 4 + 4 ) = 0.9777 = -2.23 %
So your first solution is correct.
 
  • Like
Likes Dullard
  • #3
(Quoting seems not to work...)
"The voltage across the bridge is reduced by the lead resistance, which reduces the excitation current."
Both are true, but not related as such. The higher overall resistance (Rb + 2*Rw) reduces Ie.

"The 4 resistor bridge looks like one resistor to the excitation current, Ie."
That's the opinion of one camp, the other side claim the splitting of Ie in the bridge makes the difference.
The bridge is not one resistor, but a network.
 
  • #4
MarsBravo said:
Both are true, but not related as such. The higher overall resistance (Rb + 2*Rw) reduces Ie.
I apologise for not welcoming you top PF, and for providing too much information.

MarsBravo said:
"The 4 resistor bridge looks like one resistor to the excitation current, Ie."
That's the opinion of one camp, the other side claim the splitting of Ie in the bridge makes the difference.
The bridge is not one resistor, but a network.
It is a 4 terminal network, but only two terminals are used to excite the bridge. The sensitivity of the bridge is proportional to the bridge excitation current.
 
  • #5
The supply on the input terminals do not really 'exite' the bridge, but the strain on the sensors brings the bridge out of balance and can be measured on the output terminals. The supply is a passive voltage source.
But the question remains: is the Wheatstone bridge network exactly the same as an equivalent resistor with small cable resistors in series with the shared voltage supply?
Looking only from the outside supply world, the bridge can be considered as an eqivalent resistor (first option), but inside the bridge this supply current Ie is split in two, so 'seen' from the strain sensors these cable resistors can be considered as of double value (Kirchhof, second option).
This dispute is amongst several well seasoned engineers (MSc, BSc) at our work, ranging from 20 to 40 years of experience in electronics. The formula of the first option is in use for over twenty years, but network analysis seems to suggest the obvious understanding as wrong. From the traditional outside exitation viewpoint there is no doubt, but from another angle, the inside bridge viewpoint, each strain sensor 'sees' a double cable resistor value. Hence the second option. The key here is to understand what the splitting of the exitation current will mean really. Hammering down on the tradition only sounds like a one hand clapping machine.
What's 'top PF' btw?
 
  • #6
Welcome top PF.

If you view the bridge as two branches, you do not change the differential output, but you must remember it is there.

MarsBravo said:
The supply on the input terminals do not really 'exite' the bridge, but the strain on the sensors brings the bridge out of balance and can be measured on the output terminals. The supply is a passive voltage source.
You invented the word "exite", I used the word "excite".

To quote, highlight the text, then a box appears, select either 'add to quotes', or 'reply with quote'.

I refer to the electrical excitation current, that results in a differential output voltage when the four gauges are strained. Your passive voltage source is not at the bridge. Perform an experiment, insert a 330 ohm resistor into each line from the voltage source. How much does that change the sensitivity?

Nowhere do you actually define the orientation polarity of the gauges used to form the bridge. You must verify that the assumed arrangement is actually being used in practice. If one gauge is rotated, half the signal will be lost, and the experiment will be wasted.

Your argumentative nature will keep this going, so long as you can find people silly enough to argue against the results of a practical experiment, or the results of a SPICE simulation. Do you also argue that the world is flat?
 
  • #7
Hi @MarsBravo and
..:welcome:

Probably the easiest way to settle your problem is to visit the website of the strain gauge manufacturer and look for some application notes that explain the usage details with nice drawings.

Cheers,
Tom
 
  • #8
Here is an LTspice model of the 350Ω bridge with lead resistance of 4Ω.
Resistance change is ±1%.
Strain_Gauge_Bridge.png

The model shows that it does not matter how you conceptually model the currents in the wires.

For the same resistive change of the bridge;
Without lead resistance, the output is 9.95000 mV;
With lead resistance, the output is 9.72767 mV;
The ratio is 9.72767 / 9.95000 = 0.977655
1 – 0.977655 = 0.022345 = 2.2345%

That agrees well with my original calculation in post #2;
1 - (350 / ( 350 + 4 + 4 ) ) = 0.022346 = 2.2346%
 
Last edited:
  • Like
Likes Dullard
  • #9
Post #2 answers your question. Post #8 drives a stake through it's heart.

I would (humbly) add:

Because the instrumentation amplifier is 'high impedance,' the resistance of those wires may be ignored.

People who actually do this (strain gauge instrumentation) tend to use 'shunt cals' to characterize the error due to cable resistance. If you read up on the details of that, the answer provided here may make more sense to you.
 

FAQ: Wheatstone bridge with four active strain gauges - extension cable length compensation formula

What is a Wheatstone bridge with four active strain gauges?

A Wheatstone bridge with four active strain gauges is a circuit configuration used to measure strain by converting mechanical deformation into a change in electrical resistance. All four arms of the bridge are active strain gauges, which helps in improving the sensitivity and accuracy of the measurements.

Why is extension cable length compensation necessary in a Wheatstone bridge circuit?

Extension cable length compensation is necessary because the resistance of the cables connecting the strain gauges to the measurement device can affect the accuracy of the readings. Variations in cable length or temperature can introduce errors, so compensation formulas are used to correct for these influences and ensure accurate strain measurements.

What is the formula for compensating the extension cable length in a Wheatstone bridge with four active strain gauges?

The formula for compensating the extension cable length typically involves accounting for the additional resistance introduced by the cables. One common approach is to measure the resistance of the cables separately and then subtract this value from the total measured resistance. The exact formula can vary depending on the specific configuration and the properties of the cables used.

How does temperature affect the resistance of the extension cables in a Wheatstone bridge circuit?

Temperature changes can cause the resistance of the extension cables to vary, which in turn affects the overall resistance measured by the Wheatstone bridge. This can lead to errors in strain measurement. Temperature compensation techniques, such as using temperature sensors or selecting cables with low temperature coefficients, are often employed to mitigate this effect.

Can software be used to compensate for extension cable length in a Wheatstone bridge setup?

Yes, software can be used to compensate for extension cable length in a Wheatstone bridge setup. Modern data acquisition systems often include features for inputting cable resistance values or temperature coefficients, which can then be used to automatically adjust the measurements. This makes it easier to obtain accurate strain readings without manual calculations.