Why are the S11 and S22 dips different in my microstrip resonator measurement?

  • Thread starter uWave_Matt
  • Start date
  • Tags
    Variation
In summary, the resonator has a Q-of around 10000 which is much higher than expected. It's possible that the resonator was not fabricated correctly or that there is a problem with the VNA. It's also possible that the resonator's Q is due to the HTSC material it's made out of.
  • #1
uWave_Matt
21
0
I've fabricated a simple half-wave microstrip resonator with gap-coupled microstrip feed lines. It's designed to operate around 10GHz, but due to slight over etching, the half-wave portion is 5-10um shorter than I intended. What might be more significant is that the gap, which was originally designed to be 5um, is now closer to 10um.

I'm not sure if any of the aforementioned is relevant to my question...when I characterize the device on the "trusty" HP8510, my S22 dip (9.7GHz) occurs at a slightly lower frequency than my S11 dip (10.5GHz) and has a significantly greater (more negative) magnitude.

So far, I've double checked all of my calibrations.

I suspect feed line length (where my contact touch the feed line), coupling gap variation, or perhaps a problem with the 8510. Does anyone have any other thoughts?

I've posted this question on another forum and it was suggested that I use "stepped" mode instead of "swept" mode. Does anyone know where these settings are?
 
Engineering news on Phys.org
  • #2
If the resonator is symmetric S11 should be indentical to S22, meaning your result doesn't make physical sense which in turn would point to a problem with the VNA.
Have you tried switching ports? I.e. connected port 1 to "port B" of the resonator and vice versa. The result of a reflection measurement should obviously not depend on which port of the VNA you use.

Also, it can't be a calibration issue as such; when you calibrate a VNA you calibrate the magnitude and not the frequency (although if the calibration if way off the position of the peak might appear to move; simply becaue the magnitude is wrong).

What is the Q (roughly) of your resonator?
 
  • #3
f95toli said:
If the resonator is symmetric S11 should be indentical to S22, meaning your result doesn't make physical sense which in turn would point to a problem with the VNA.
Have you tried switching ports? I.e. connected port 1 to "port B" of the resonator and vice versa. The result of a reflection measurement should obviously not depend on which port of the VNA you use.

I've rotated the jig that the resonator is in which effectively causes S11 and S22 to switch. This is just the room temperature measurement. I'm dealing with HTSC material so the S-parameters should be significantly different after I cool the chamber.


f95toli said:
Also, it can't be a calibration issue as such; when you calibrate a VNA you calibrate the magnitude and not the frequency (although if the calibration if way off the position of the peak might appear to move; simply becaue the magnitude is wrong).

But you have to calibrate the magnitude over a certain frequency range.

f95toli said:
What is the Q (roughly) of your resonator?

Since it's an HTSC, the Q-should be just about humongous. Another HTSC resonator fabricated in this lab had Q>10000 and I've calculated the Q of this one to be close to that.

I should also mention that one of the coaxial lines from the VNA to the jig is 4-5 time longer than the other, although I'd assume calibration would take care of this.
 

FAQ: Why are the S11 and S22 dips different in my microstrip resonator measurement?

What is the difference between S11 and S22 in terms of variation?

S11 and S22 are both scattering parameters that measure the amount of reflection and transmission of an electromagnetic wave as it passes through a circuit. The main difference between the two is their placement in the circuit. S11 measures the reflection at the input port, while S22 measures the reflection at the output port.

How does the variation between S11 and S22 affect circuit performance?

The variation between S11 and S22 can indicate the level of mismatch between the input and output ports of a circuit. This mismatch can lead to signal loss and affect the overall performance of the circuit.

What causes variation between S11 and S22?

The variation between S11 and S22 can be caused by several factors, including impedance mismatches, signal reflections, and parasitic elements in the circuit. It can also be influenced by the physical layout and design of the circuit components.

How can the variation between S11 and S22 be measured?

The variation between S11 and S22 can be measured using a network analyzer, which sends a signal through the circuit and measures the reflected and transmitted signals at the input and output ports. The resulting S11 and S22 values can then be compared to determine the level of variation.

Can the variation between S11 and S22 be reduced?

Yes, the variation between S11 and S22 can be reduced by designing the circuit with proper impedance matching and by using components with low parasitic effects. It is also important to carefully layout the circuit to minimize signal reflections and interference.