Why Don't GM Counters Detect Alpha Particles?

In summary, the conversation discusses the detection of alpha particles in GM counters and the reasons why some GM counters may not be able to detect them. It also touches on the detection of neutrons and the different methods used to detect them. A datasheet for a GM tube is referenced and the concept of energy discrimination is explained. The conversation also delves into the physics behind the detection of gamma rays and neutrons.
  • #1
Vilius
3
1
Hi,
I am an undergraduate electronics engineer building a GM counter for my final assessment. I read a lot of theory about GM counters as well as the nuclear physics theories in general, but there is one thing I can not find an answer to. Why do none of the GM counters detect alpha particles? I have looked through so many different models (both pancake and end-window types) and I did not find a single one capable of detecting alpha particles. I know that alpha has the smallest penetrating capability, but at the same time, it has the largest charge and inertia as well (due to a comparably highest mass of all radioactive particles.) Thank you in advance for any observations.
 
Engineering news on Phys.org
  • #2
Any GM tube with a thin window (usually mica) should detect alpha particles. You will find there will be a minimum energy and that will depend on how thick the window is. A good tube to look at is the LND712. It's small and typically quite expensive, but it's the classic. Anything with a similar window thickness around 2mg/sqcm should do the same job.
 
  • Like
  • Informative
Likes mark_bose, berkeman, Lord Jestocost and 2 others
  • #3
Vilius said:
Why do none of the GM counters detect alpha particles?
'None' is a bit of a stretch (as it was already stated above), but in general, most GM detectors (and the 'most' here comes from the fact that most of them are made cheap) does not have the right type of GM tube and cover to provide adequate window for them.

Though there are other detectors which does not have window at all:wink:

 
  • Like
Likes Alex A
  • #4
And what about neutrons? A popular explanation is that the neutrons only have mass with no charge, so they can not ionize the gas inside a GM tube, but gamma photons also do not have a charge, however, gamma rays are detected...
And the last thing. GM counter datasheets often provide ,,maximum detectable beta particle /gamma ray energy" I can understand, why low energy particles/rays do not get detected, but what is the physics behind maximum detectable energy?

Thank you in advance
 
  • #5
Can you point to a datasheet with a maximum energy limit? On a proper sheet you will get a response curve, which may tail off at high energies, because these interact less with a given amount of matter.

Gamma and X-ray photons are not charged, but they interact by the Photoelectric effect or the Compton effect, and so produce high energy electrons which are charged and ionising - they leave a trail of charged pairs in their wake. A GM tube may only detect 1% of the gamma rays that go through it depending on the energy and construction, and making the metal shell thicker usually results in a tube that is a better gamma detector.

Neutrons are a similar case in that they can for example knock protons out of a hydrogen containing material which is then an ionising radiation and can be detected by a scintillator or a GM tube. This is how neutrons were discovered - a block of paraffin placed in front of a GM tube increased the counts which suggested something previously invisible was being detected. Slow neutrons are often detected by having materials that produce nuclear fragments of high energy, and usually in tubes that operate below the Geiger plateau (and so the pulse size depends on the number of ion pairs produced, a GM tube produces the same large pulse for every event). Nuclear fragments tend to drop a lot of energy in a small path length, producing a big pulse. This enables energy discrimination, which is important, because you want to know what you are detecting. Neutron detection is a huge field and there are radically different ways of doing this, including weirdly simple ways like wrapping silver foil around a GM tube. All methods have strengths and weaknesses.
 
  • #6
I am using an obsolete Philips 18504 or ZP1400 (one tube has 2 names, really strange, but it is what it is...)
So here is a picture from the datasheet.

,,A GM tube may only detect 1% of the gamma rays that go through it depending on the energy and construction, and making the metal shell thicker usually results in a tube that is a better gamma detector."

Can you explain the reason for that?

About the neutrons, does it mean that neutrons can only ionise hydrogen?

It started as a simple question, but now I see that it went really far... I am really curious :)

Thank you again
 

Attachments

  • 1.jpg
    1.jpg
    46.9 KB · Views: 118
  • Like
Likes Alex A
  • #7
Vilius said:
And what about neutrons? A popular explanation is that the neutrons only have mass with no charge, so they can not ionize the gas inside a GM tube, but gamma photons also do not have a charge, however, gamma rays are detected..
Neutrons would be detected in a gas, e.g, 10BF3 or 3He, where the 10B and 3He absorb a neutron and fission, for former undergoing an (n,α) reaction and the latter, an (n,p) reaction. Also, depending on the neutron energy, one might surround the tube with paraffin or other hydrogenous material to thermalize the neutrons.

Edit/update: One could also coated a Geiger counter or proportional counter tube with a strong neutron absorber, e.g, Hf, or B4C, with B enriched in 10B, or HfB2, or even 235U. The last option is a fission chamber, which uses fission reactions to interact with neutrons.

Otherwise, one might use so-called Self-Powered Neutron Detectors
https://www.oecd-nea.org/science/rsd/ic96/4-2.pdf
 
Last edited:
  • Informative
  • Like
Likes mark_bose and Alex A
  • #8
For a detector to be useful, the important question is not how many events it is detecting, or missing, but how many are out there and how accurately that can be estimated. That table gives information on what the tube can and can't detect and how much you would need to adjust the numbers to get the true answer.

GM tubes are usually made of thin materials so to be sensitive to beta rays etc. Gamma rays don't interact well, especially at high energy, so most will sail through the tube like it isn't there.

Hydrogen isn't a great way to detect neutrons (people that work with proton recoil counters may disagree), and putting wax in front of a GM tube with a window is highly inefficient, but a proton and a neutron have similar masses so when a neutron does hit a proton the maximum kinetic energy transfer can happen. If you are going to play marbles, you want it to hit another marble. A bowling ball would probably barely move, and you wouldn't expect a GM tube to detect neutrons normally.
 
  • Like
Likes mark_bose and Astronuc

Related to Why Don't GM Counters Detect Alpha Particles?

What is a GM counter and how does it work?

A GM (Geiger-Müller) counter is a type of radiation detection device that measures ionizing radiation levels. It consists of a Geiger-Müller tube filled with an inert gas like helium, neon, or argon at low pressure. When ionizing radiation, such as alpha particles, enters the tube, it ionizes the gas, causing a momentary conductive path that generates a measurable electrical pulse. This pulse is then counted and displayed, allowing users to quantify the radiation level.

Can a GM counter detect alpha particles?

Yes, a GM counter can detect alpha particles, but its efficiency depends on the design of the counter. Alpha particles have a very short range in air and can be stopped by a few centimeters of air or a thin barrier. Therefore, a GM counter designed to detect alpha particles typically has a thin window, often made of mica, that allows the alpha particles to enter the tube and ionize the gas inside.

What are the limitations of using a GM counter for alpha particle detection?

The primary limitation of using a GM counter for alpha particle detection is the need for a thin entrance window, which makes the device more fragile and susceptible to damage. Additionally, alpha particles have low penetration power, so the counter must be very close to the source of radiation. The GM counter also cannot differentiate between different types of radiation (alpha, beta, gamma) without additional shielding or discrimination techniques.

How can you ensure accurate measurement of alpha particles with a GM counter?

To ensure accurate measurement of alpha particles with a GM counter, the device should have a thin window to allow alpha particles to enter. The counter should be placed as close as possible to the radiation source, ideally within a few millimeters. It's also important to calibrate the GM counter regularly and ensure it is in good working condition. In some cases, using a specialized alpha probe attachment can enhance the accuracy of measurements.

What are the safety precautions when using a GM counter to detect alpha particles?

When using a GM counter to detect alpha particles, safety precautions include avoiding direct contact with radioactive sources and using tools to handle them. Ensure the GM counter's thin window is not damaged, as this can compromise its ability to detect alpha particles. Always follow proper radiation safety protocols, such as wearing protective clothing and using shielding when necessary. Additionally, work in a well-ventilated area to prevent the inhalation of any radioactive particles that might be released.

Similar threads

  • Nuclear Engineering
Replies
20
Views
4K
  • Introductory Physics Homework Help
Replies
12
Views
563
  • High Energy, Nuclear, Particle Physics
Replies
6
Views
2K
  • Introductory Physics Homework Help
Replies
1
Views
1K
  • Introductory Physics Homework Help
Replies
1
Views
1K
  • High Energy, Nuclear, Particle Physics
Replies
8
Views
3K
Replies
4
Views
1K
  • Introductory Physics Homework Help
Replies
2
Views
1K
  • Biology and Chemistry Homework Help
Replies
2
Views
2K
Replies
9
Views
2K
Back
Top