X-ray fluorescent screens and film contrast

  • Thread starter ChumpusRex
  • Start date
  • Tags
    Film X-ray
In summary: In addition, at the high exposure end, the response of the film starts to saturate as graininess becomes more pronounced.In summary, using a fluorescent screen in radiography can increase the overall sensitivity and contrast of the system. However, this increase in contrast is due to non-linear processes in the screen and film, caused by factors such as emulsion composition, particle size and shape, and screen composition. Newer rare Earth compounds have replaced older materials such as Calcium Tungstate and Gadolinium oxysulphide in screen materials.
  • #1
ChumpusRex
8
0
It's common practice in radiography, especially medical radiography where radiation doses are relevant, to use a fluorescent screen to expose photographic film, rather than rely on the effect of x-rays on the film itself.

I understand the concept of fluorescence and the overall photon gain/quantum efficiency gain of using a heavy metal fluorescent material (e.g. Calcium Tungstate or Gadolinium oxysulphide) to absorb the x-rays photons and emit numerous visible light photons. This subsequently increases the overall sensitivity (speed) of the system.

However, a fluorescent screen is also stated to increase the film contrast (gamma) significantly.

As gamma is a log-log measurement this would suggest that there is a non-linear process in the screen. As surely, if the conversion was linear (e.g. 100 light photons for 1 x-ray photon) then the system gamma would be unaffected.

Does anyone know what this non-linearity is, and what factors affect it?

I've already been through about half-a-dozen radiography texts, all of which mention the effect on gamma, but none of which explain it.
 
Science news on Phys.org
  • #2
I have to give just a partial answer here, because all my texts on the matter are at the office (which is where I'm not at the moment). Most of what I put here is recollection from a fuzzy memory. It's been a while since I last worked with film/screen...most of what I do now is on CR.

There are lots of non-linearities in film or film/screen systems. They come from things like the emulsion composition, emulsion particle shapes and sizes, the screen composition and grain to name a few off the top of my head.

Film makers tailor the film gamma and shape of the H&D curve by changing the shape and size of the emulsion crystals. They can also do the same with the screen.

Also, the conversion process in the screen is not entirely linear either. Keep in mind that your conventional x-ray beam is polychromatic, and x-ray photons of different energies will produce different numbers of light photons (which is a statistical process).

For more details on film and film/screen in medical imaging, I highly recommend the two following books by Art Haus

Basics of Film Processing in Medical Imaging
Advances in Film Processing Systems Technology and Quality Control in Medical Imaging

BTW, Calcium Tungstate or Gadolinium oxysulphide have been replaced long ago as screen materials by rare Earth compounds.

If you're still interested by tomorrow, I'll look up some more details and post a hopefully better reply (unless someone beats me to it)
 
  • #3
imabug said:
BTW, Calcium Tungstate or Gadolinium oxysulphide have been replaced long ago as screen materials by rare Earth compounds.

duh, so used to CR phosphors, I'm forgetting screen/film things. Gadolinium oxysulphide is one of the newer rare Earth phosphors.
 
  • #4
"The Physics of Radiology" by Johns and Cunningham has this to say about differences between H&D curves for film and film/screen:
The "no screen" film shows a linear increase in density with exposure up to a density of about 1.0 and then gradually starts to saturate. In contrast, the response of the film used with a screen is sigmoidal. The difference can be explained as follows. The film without screen is exposed by x-ray photons, each of which deposits enough energy to sensitize a single grain of silver, hence the initial linear relation; as the silver grains are used up the curve starts to show saturation.

When films are exposed with screens the x-ray photons are absorbed in the screens, which emit visible and ultraviolet light. It is these light photons that cause 99% of the exposure to the film. Since it requires the absorption of several of these light photons to sensitize a single grain of silver, we obtain a sigmoidal response at small densities.
 

FAQ: X-ray fluorescent screens and film contrast

What is the purpose of X-ray fluorescent screens?

X-ray fluorescent screens are used to convert X-ray energy into visible light, which allows for the detection and visualization of X-ray images.

How do X-ray fluorescent screens improve image contrast?

X-ray fluorescent screens contain elements that emit light when exposed to X-rays, resulting in a brighter image with increased contrast compared to traditional X-ray films.

How does film contrast affect X-ray image quality?

Film contrast refers to the range of densities present on an X-ray film, which directly impacts the image quality. Higher contrast images have a greater difference between light and dark areas, making it easier to distinguish between structures.

What factors can affect the contrast of an X-ray image?

The contrast of an X-ray image can be influenced by factors such as the X-ray machine settings, type of X-ray film used, and the thickness and composition of the object being imaged.

Are there any limitations to using X-ray fluorescent screens and film?

X-ray fluorescent screens and film are limited in their ability to produce high-resolution images and can be affected by ambient light and chemical processing errors. They also require regular maintenance and calibration to ensure accurate and consistent results.