MRI Safety: Question on Potential Burns

  • Thread starter Thread starter vabamyyr
  • Start date Start date
  • Tags Tags
    Mri Safety
AI Thread Summary
MRI safety protocols require the removal of unplugged coils from the bore during scans to prevent patient burns and image quality degradation. Unplugged coils can create a conducting loop with the patient and table, potentially inducing dangerous currents in the presence of RF fields. High voltage can also be induced on open connectors due to patient movement or switching the magnetic field on and off, posing a risk if touched. Coils designed to be resonant at the system's frequency can absorb significant energy during RF transmission, leading to overheating if not properly connected. Properly plugging in coils allows for detuning, minimizing energy absorption and maintaining image quality.
vabamyyr
Messages
65
Reaction score
0
Hi,

MRI safety manual says that whenever you are scanning a patient with a certain surface coil you should remove all unplugged coils from the bore during the scan. For example, if you have a head/neck + spine coil matrix with two coil ports and you are scanning only the spine part then you should either connect the head part also or remove head/neck coil from the bore.
I was wondering what is the physical mechanism that can cause patient burns?
For example, if you scan patient's spine and leave the head coil unplugged. I was thinking in the lines that somehow when the patient is in contact with the unplugged coil it could create a conducting loop between the unplugged coil, patient, patient table maybe and when this loop is in a RF field it causes current to flow in that loop that could cause burns to the patient.
 
Physics news on Phys.org
Thats not burns, but high voltage shock, what is dangerous.

MRI involves very strong magnetic field, so high voltage could get induced on open connectors of the coil as it changes its position with even small patient's moves or when the magnetic field is switched on/off. That voltage could possibly be dangerous if someone touches the connector. The same coil could also catch RF magnetic fields of other coils and thus got inducted voltage.
 
Last edited:
I think the concern would be two-fold, burns and bad images. A MRI coil is a structure which is, by design, resonant at the system's frequency. So, when you are running a pulse sequence that deposits a lot of RF then the coil will be absorbing a lot of energy, perhaps a few kilowatts, and could get hot. Plugging the coil into the table allows the system to "detune" the coil so that it is no longer resonant when the system is transmitting RF. Since it is not resonant, it absorbs little energy, perhaps a few milliwatts, and does not get hot.

With the interference of a resonant structure, even if it doesn't get hot, it should degrade your image quality, i.e. change local flip angles, add noise by inductive coupling to a bigger volume, etc.
 
Thank you DaleSpam for your reply.
 
The administrator at this forum (Bertus) is an MRI specialist who might answer your question authoritatively.

www.allaboutcircuits.com

go well
 
Assume that this is a case where by sheer coincidence, two sources of coherent single-frequency EM wave pulses with equal duration are both fired in opposing directions, with both carrying the same frequency and amplitude and orientation. These two waves meet head-on while moving in opposing directions, and their phases are precisely offset by 180 degrees so that each trough of one wave meets with the crest of the other. This should be true for both the electric and magnetic components of...